Hi,
Just received NOS 1999 densitometer - never used, with invoices, manual, box, etc. There were no calibration strips included, but I checked it against Stouffer T2115 wedge and print density reference tile provided within RH Designs calibration kit. Here are the results:
1. 0.00
2. 0.12
3. 0.28
4. 0.44
5. 0.58
6. 0.72
7. 0.88
8. 1.01
9. 1.14
10. 1.29
11. 1.44
12. 1.59
13. 1.74
14. 1.90
15. 2.04
16. 2.18
17. 2.33
18. 2.47
19. 2.61
20. 2.77
21. 2.93
Paper Base - 0.00
D=0.04 - 0.07
D=90% - 1.95
D.max - 2.25
Do you think the results are OK? In case of transmission they look quite consistent and precise, but I'm worried about reflection - if I zeroed the meter on paper base tile, I should get exactly 0.04 on the 0.04 tile, but I got 0.07 - more than 0.02 tolerance mentioned in the specs in the manual.
Just received NOS 1999 densitometer - never used, with invoices, manual, box, etc. There were no calibration strips included, but I checked it against Stouffer T2115 wedge and print density reference tile provided within RH Designs calibration kit. Here are the results:
1. 0.00
2. 0.12
3. 0.28
4. 0.44
5. 0.58
6. 0.72
7. 0.88
8. 1.01
9. 1.14
10. 1.29
11. 1.44
12. 1.59
13. 1.74
14. 1.90
15. 2.04
16. 2.18
17. 2.33
18. 2.47
19. 2.61
20. 2.77
21. 2.93
Paper Base - 0.00
D=0.04 - 0.07
D=90% - 1.95
D.max - 2.25
Do you think the results are OK? In case of transmission they look quite consistent and precise, but I'm worried about reflection - if I zeroed the meter on paper base tile, I should get exactly 0.04 on the 0.04 tile, but I got 0.07 - more than 0.02 tolerance mentioned in the specs in the manual.