    Densitometer Calibration

    Can someone walk me through the calibration procedure for an x-rite 810?

    I have a manual but, I'm a little thick these day and need some hand holding.


    Thane in advance.

    Joe

  2. #2

    Re: Densitometer Calibration

    Joe,
    It has been a while for me, but as I recall, you just follow the direction in the manual. Can you be more specific?

  3. #3

    Join Date
    Mar 2002
    Location
    The "Live Free or Die" state
    Posts
    521

    Re: Densitometer Calibration

    Do you have the calibration negative (transmissive) and plate (reflective)? If not you won't be able to calibrate following the books directions. If you do, the book explains it better than I can.

  4. #4
    Hack Pawlowski6132's Avatar
    Join Date
    Sep 2009
    Posts
    769

    Re: Densitometer Calibration

    I have the plate and the negative. However, the negative is pretty damaged and I woudn't trust it. I do, however, have something I ordred from Stouffer last year. It's a 4x5 5 Step Calibrated Wedge. Can I use that? If I'm only using this for B/W film testing, do I still need to calibrate for reflective measurements?

  5. #5

    Join Date
    Jul 2007
    Location
    Austin TX
    Posts
    1,689

    Re: Densitometer Calibration

    Yes you can use a 5 step calibrated wedge and interpolate for in between densities providing that wedge covers your range of interest. Make a plot of the Stouffer wedge densities vs the densitometer readings. Forget about reflective measurements if you are not needing that mode of operation.

    Do your calibration run in a darkened room to avoid reflected room light from influencing your readings, especially when checking high densities.

    Nate Potter, Austin TX.

  6. #6

    Join Date
    Apr 2006
    Location
    Hell's Kitchen, New York
    Posts
    528

    Re: Densitometer Calibration

    What is the highest density step on the Stouffer wedge you have? (4.0?) After pressing CAL twice to enter calibration/setup mode you should be able to set that as the CAL HI value in all four channels - assuming that it is neutral, and assuming that you do not have have Status M red, green and blue values. Then complete the zero and CAL HI. Just follow the exact sequence of button presses listed in the manual.

    Unless you have a calibrated target your measurements will not be bang-on accurate (ie the truly correct absolute value), but that may not matter. Follow the same calibration procedure every time and you should have precise enough measurements for comparative purposes, albeit without great accuracy.

    Good luck,
    Helen

Similar Threads

  1. Macbeth densitometer help?
    By jvuokko in forum Darkroom: Film, Processing & Printing
    Replies: 0
    Last Post: 28-Dec-2008, 15:28
  2. Noritsu DM-1 Densitometer Calibration
    By Neil Miller in forum Style & Technique
    Replies: 7
    Last Post: 12-Mar-2008, 05:50
  3. How does a"scanning" densitometer differ from a regular densitometer?
    By octagon in forum Darkroom: Film, Processing & Printing
    Replies: 1
    Last Post: 14-May-2002, 20:16

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •