    Densitometer reading upside-down?

    I have a Konica "nice print system" densitometer I bought on eBay. It was reading ok for a while, but it was a little off (compared to a calibrated step wedge.) I noticed that there was some dust in the reading area, so I blew it out with some dust-off. Now I get strange readings... 2 minus the reading gives me the correct reading. It also has a minus sign in front , but I'm not sure if it was there before.

    For instance I read
    -1.96 for a wedge that's calibrated valuel is .04
    -1.83 = .17
    -1.67 = .33
    etc.

    Does anyone know this densitometer? I guess I could use it like this, but it's a pain.

  2. #2

    Join Date
    Sep 2003
    Location
    Portland, OR
    Posts
    743

    Re: Densitometer reading upside-down?

    I don't know about your meter - maybe you could post a photo of it as several different meters have been rebranded, like Noritsus and X-rites I've seen.

    Anyway, it looks like your meter was rezeroed somehow and it now has a zero that is negative. So everytime you take a reading of something, which has a positive density value, you get the density number added to the negative number. I hope that made sense. i.e.
    (-1.96 + 0.33 = -1.67)

    I suggest that you rezero it and see if that helps.

  3. #3

    Join Date
    Nov 2007
    Location
    Nuremberg Germany
    Posts
    1,048

    Re: Densitometer reading upside-down?

    When the offset of the amplifier is misadjusted, you will get negative readings. So adjust the densitometer properly.

    Peter K

  4. #4

    Join Date
    Apr 2006
    Location
    Hell's Kitchen, New York
    Posts
    525

    Re: Densitometer reading upside-down?

    In the very unlikely event that you don't already know how to do this: with no film in place take a reading and adjust the zero calibration to zero. No calibration targets are required for setting zero. Then you can use your wedge to calibrate the span (ie the higher end of the density range).

    Best,
    Helen

  5. #5

    Join Date
    Dec 2007
    Location
    Huntington, Long Island, NY
    Posts
    167

    Re: Densitometer reading upside-down?

    I do know that. I was trying to get it to work, and discovered that if I zeroed it out on the first step of the step wedge, then I get correct readings, but offset by the density of the first step. If I...

    zero it out on the .04 wedge
    then the second wedge is .13 (.17 - .04)
    the third one is .29 (.33 - .04)

    but if I zero it out with no film in the reader, I get the negative numbers.

    Strange.

  6. #6

    Join Date
    Apr 2006
    Location
    Hell's Kitchen, New York
    Posts
    525

    Re: Densitometer reading upside-down?

    Very strange. I Googled 'konica nice print system densitometer' and found this thread which mentions its similarity to an XRite 320.

    The negative readings imply that the receptor is getting more light back than when it was zeroed, and one way for that to happen would be if a reflection head was zeroed with nothing in place. That seems partially consistent with some of your observations, but not all. Of course you can't zero a reflection densitometer.

    If you zero it with the 0.04 step in position, what reading do you get with nothing in position?

    Best,
    Helen

  7. #7

    Join Date
    Dec 2007
    Location
    Huntington, Long Island, NY
    Posts
    167

    Re: Densitometer reading upside-down?

    Good question. I'll try that tonight...

  8. #8

    Join Date
    Nov 2007
    Location
    Nuremberg Germany
    Posts
    1,048

    Re: Densitometer reading upside-down?

    There is something wrong with the electronic, possible a broken OP-amplifier. Heat up your soldering iron.

    Peter K

  9. #9

    Join Date
    Dec 2007
    Location
    Huntington, Long Island, NY
    Posts
    167

    Re: Densitometer reading upside-down?

    When I zero it on the .04 strip, then reading nothing returns -1.96

  10. #10

    Join Date
    Aug 2000
    Location
    California
    Posts
    3,908

    Re: Densitometer reading upside-down?

    I am beginning to wonder if this densitometer reads both transmission and reflection.
    If so, it may be set for reflective while attempting to read transmission.
    Also, it is possible to zero a reflection densitometer with the correct calibration tablet.

Similar Threads

  1. Scanner as densitometer?
    By false_Aesthetic in forum Digital Hardware
    Replies: 9
    Last Post: 15-Mar-2016, 10:02
  2. Noritsu DM-1 Densitometer Calibration
    By Neil Miller in forum Style & Technique
    Replies: 7
    Last Post: 12-Mar-2008, 05:50
  3. Got Stoffer 21 Step Tablet, Densitometer? Pls Help
    By Andre Noble in forum Darkroom: Film, Processing & Printing
    Replies: 3
    Last Post: 30-Dec-2004, 07:50
  4. How does a"scanning" densitometer differ from a regular densitometer?
    By octagon in forum Darkroom: Film, Processing & Printing
    Replies: 1
    Last Post: 14-May-2002, 20:16
  5. Densitometer for B&W Negative
    By d.s. in forum Style & Technique
    Replies: 3
    Last Post: 5-Feb-2002, 07:50

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •